Operation
6
Drying
Maintenance,
7
inspection and
testing of the
accessories
8
Packaging
9
Sterilization
10
Storage
Instructions
disinfectant solution
combined with
proteolytic enzyme if
possible.
- Rinse well the
accessories in flowing
water
No particular
requirements
No particular
requirements
Pack the accessories in
sterilization pouches
- Steam sterilization at
135°C (275°F) during
10 minutes in gravity
type autoclave
(Tabletop, N type)
- Drying time after
sterilization – 30
minutes
Keep sterilized
accessories in
sterilization pouches in
a dry and clean
environment
- 11 -
Details and Warnings
(concentration, immersion time,
etc.)
- Do not use disinfectant solution
containing aldehyde, phenol or any
products which may damage the
accessories
- Check the validity period of the
pouch given by the manufacturer
to determine the shelf life of the
sterilized items
- Use packaging which is resistant
up to a temperature of 141°C
(286°F)
- Follow maintenance and operation
procedures of the autoclave
provided by the manufacturer
- The only sterilization parameters
to be used are those that have
been validated and provided to
the user in this User Manual
Sterility cannot be guaranteed if
packaging is open or damaged
(check the packaging before using
the accessories)