• For your own safety, please use personal protection gear (gloves, mask).
5.
Adverse Reactions
None.
6.
Step-by-Step Instructions
6.1
Cable connection test
Connection test feature is included in Wirele-X in order to check the cables:
6.1.1
Insert
attached Lip Clip and File Clip into the
device receptacle.
6.1.2
Verify that the File Clip contact and the Lip
Clip are clean and make connection
between them.
6.1.3
LEDs of all bars will turn on indicating
proper connection of cables (Pic. 1).
6.1.4
If LEDs on the bars does not appear, the
Measurement cable or the File Clip should
be replaced.
6.2
Getting Started
Prior the first use, it is recommended to sterilize the Lip Clip, the File Clip and the Touch
Probe. Please refer to section 8 for further information regarding Wirele-X maintenance.
6.2.1
Disconnect the charger from the device if
connected and turn the device on by
pressing the On/Off button. Blinking blue
color LEDs around the On/Off button
indicating the device is On (Pic. 2).
6.2.2
Before connecting the Measurement cable
with attached Lip Clip and File Clip to the
patient, plug the Measurement cable into
the device receptacle.
6.2.3
Attach the Lip Clip to the patient lip.
the
Measurement
Lip Clip
cable
with
- 5 -
Pic. 1
Pic. 2